In a message dated 5/7/07 3:13:00 AM, eeshdewan@... writes:
> When we have calibrated our display to gamma 2.2, 6500K, white luminance
> 90cd/m2 and black luminance to 0.3cd/m2 and then open the colorimeter window to
> measure the color temperature of white, the reading show 5500 odd Kelvin and
> not 6500k. Is there something that I am doing wrong?
>
It sounds like the colorimeter window is not measuring your calibrated
whitepoint, but your native whitepoint. Calibrate your monitor, go to the Before and
After screen at the end of the software, open the colorimeter window, and
check the K value in both before, and in after, mode.
Keep in mind that Kelvin values are highly leveraged, and that a variation
of hundreds of K points can be as little as .003 difference in measured xyY
values. This means that end user spectros and colorimeters (and to a suprising
degree, even laboratory devices) are not capable of working down to the 100k
level, as large a number as that sounds. Frankly, K measurements should be
reported to the nearest 500K, not the nearest 1K, to reflect the actual value of
the measurements. 6 and a half thousand K as the next unit from 6 thousand would
be more reasonable units of measurement. Adding all those extra units makes
end users assume that the tolerances are actually that close.
C. David Tobie
Product Technology Manager
ColorVision Business Unit
Datacolor Inc.
CDTobie@...
www.colorvision.com
**************************************